i1Pro instrument re-calibration

I was looking over my documents and found the original calibration page that was shipped with my instrument. It recommends that I should have the i1Pro re-calibrated after about 2 years. My instrument is about 5 years old now.

What is your recommendation regarding re-calibration? How can you tell if it is needed? My instrument is always stored in in the padded case except when in use.

tks, louie

Use i1Diagnostics v2.5.1 to check your i1Pro.

You can post the result log file here for diagnosis.

Download here:
xrite.com/product_overview.a … wareID=506

Here is the log file output from i1Diagnostics. Please take a look and see if there is any indication that I should recalibrate.

tks, louie

General Information
  Date and Time: 5/3/10 11:31 AM
  Application version: Version 2.5.1
  SDK version i1: Version 3.4.0  Build 131
  SDK version iO: Version 1.1.2  Build 100
  SDK version iSis: Version 1.0.5  Build 70
  Platform: Apple Mac OS X 10.5.8 (9L31a) i386

Device Information
  Device type: i1Pro (NO filter) [ Ambient Light  Emission  Reflectance  Reflectance Scan  Flash ]
  Serial number: 125689
  Firmware version: 302
  CPLD version: 301
  Dark measurement count: 7536
  Lamp burning time: 7243.59 seconds
  Spot measurement count: 617
  Scan measurement count: 1312

Dongle Test
  Module "Scanner" is licensed
  Module "Monitor" is licensed
  Module "Printer" is not licensed
  Module "Printer RGB only" is licensed
  Module "Printer CMYK only" is not licensed
  Module "Beamer" is licensed
  Module "Printer Easy RGB" is licensed
  Module "Printer Easy CMYK" is licensed
  Module "Digicam" is licensed
  Module "Editor" is licensed

Button Test
  Button is working normally

Calibration
  Emission calibration: Successful

Emission Test
  White measurement: Successful
  Gray measurement: Successful
  Black measurement: Successful
  Red measurement: Successful
  Green measurement: Successful
  Blue measurement: Successful
  Emission Test: Pass

Dark measurement
  Dark measurement: Pass

Noise Measurement
  Noise measurement: Pass

Lamp test
  Reflectance calibration: Successful
  Reflectance measurement: Successful
  Check measurement result: Successful

Reflectance Test
  Reflectance calibration: Successful
  Reflectance spot test: Successful
  Scanning calibration: Successful
  Reflectance scan test: Successful
  Measurement 1: Successful
  Measurement 2: Successful
  Measurement 3: Successful
  Measurement 4: Successful
  Measurement 5: Successful
  Measurement 6: Successful
  Measurement 7: Successful
  Measurement 8: Successful
  Measurement 9: Successful
  Measurement 10: Successful
  Reflectance drift test: Pass


*** PASS ***


**************************

General Information
  Date and Time: 5/3/10 11:31 AM
  Application version: Version 2.5.1
  SDK version i1: Version 3.4.0  Build 131
  SDK version iO: Version 1.1.2  Build 100
  SDK version iSis: Version 1.0.5  Build 70
  Platform: Apple Mac OS X 10.5.8 (9L31a) i386

Device Information
  Device type: i1Pro (NO filter) [ Ambient Light  Emission  Reflectance  Reflectance Scan  Flash ]
  Serial number: 125689
  Firmware version: 302
  CPLD version: 301
  Dark measurement count: 7536
  Lamp burning time: 7243.59 seconds
  Spot measurement count: 617
  Scan measurement count: 1312

Dongle Test
  Module "Scanner" is licensed
  Module "Monitor" is licensed
  Module "Printer" is not licensed
  Module "Printer RGB only" is licensed
  Module "Printer CMYK only" is not licensed
  Module "Beamer" is licensed
  Module "Printer Easy RGB" is licensed
  Module "Printer Easy CMYK" is licensed
  Module "Digicam" is licensed
  Module "Editor" is licensed

Button Test
  Button is working normally

Calibration
  Emission calibration: Successful

Emission Test
  White measurement: Successful (X=77.768 Y=82.087 Z=88.081)
  Gray measurement: Successful (X=28.928 Y=30.439 Z=34.183)
  Black measurement: Successful (X=-0.094 Y=-0.035 Z=0.047)
  Red measurement: Successful (X=46.700 Y=22.946 Z=1.134)
  Green measurement: Successful (X=15.671 Y=49.901 Z=6.860)
  Blue measurement: Successful (X=15.078 Y=8.853 Z=79.833)
  Emission Test: Pass

Dark measurement
  Dark measurement: Pass (-27.711)

Noise Measurement
  Noise measurement: Pass (3.275)

Lamp Test
  Reflectance calibration: Successful
  Reflectance measurement: Successful
  Check measurement result: Successful

Reflectance Test
  Reflectance calibration: Successful
  Reflectance spot test: Successful (LOT=7244.25 seconds, CIT=0.01836 seconds, LIT=0.0203884 seconds)
  Scanning calibration: Successful
  Reflectance scan test: Successful
  Measurement 1: Successful (L=96.413 a=-0.591 b=2.199)
  Measurement 2: Successful (L=96.380 a=-0.584 b=2.225)
  Measurement 3: Successful (L=96.352 a=-0.569 b=2.248)
  Measurement 4: Successful (L=96.300 a=-0.562 b=2.320)
  Measurement 5: Successful (L=96.302 a=-0.578 b=2.323)
  Measurement 6: Successful (L=96.303 a=-0.582 b=2.332)
  Measurement 7: Successful (L=96.302 a=-0.578 b=2.333)
  Measurement 8: Successful (L=96.305 a=-0.589 b=2.342)
  Measurement 9: Successful (L=96.282 a=-0.575 b=2.366)
  Measurement 10: Successful (L=96.297 a=-0.586 b=2.362)
  Reflectance drift test: Pass (0.061)


*** PASS ***

The numbers on your Emission Test seem to be a bit off.
In particular, it looks suspicious to have your Black measurement have negative numbers for X and Y. Gray and Green look pretty different too.

I don’t really know how to interpret the numbers, but your numbers seem to be quite different from mine:

Patrick,

I have contacted Xrite support to see what they say about the results of test. I will keep you posted.

tks, louie

Can you tell me exactly what the emissions test is. I just ran i1 diagnostics on my i1 Pro because of some strange behavior in my profiles and it failed the emission test.

An i1Pro device can measure reflective light (light bouncing off of a paper for instance) and emissive light (light emitting from a computer display). The i1Diagnostic program runs the device through its paces on all these functions. The emissive test would not include any self-illumination, so you might try it again and make sure the device is on the screen when it is supposed to be and that it is well-settled in the in the cradle when it is supposed to be.

I ended up sending in the log files. And, it looks like I have a lamp issue. Fortunately the device is still under warranty. Thank you for your quick response.